    My LASER current meter for DC tubes.

    I had to go down to 77% power to not exceed the 26mA that RECI says is needed to reach 10,000 hours.

    http://www.thingiverse.com/thing:166861
    Triumph TR-9060 with 5200 Chiller and 100 watt Z4 RECI.

  2. #2
    Now I can adjust 100% power to be 26 mA.

    http://labs.nortd.com/lasersaur/manual/laser_adjustments
    Last edited by Robert Silvers; 10-16-2013 at 6:33 PM.

  3. #3
    From post 18 in the Triumph review thread

    China tends to set up machines in many cases by the cutting ability of the tube (output beam power)rather than by testing the actual supplied current, if they get 100 watts of beam that can on occasion mean over driving the tube.(not all of them by any means but some do it)
    Looks like China did exactly that on yours Rob, Matched power supplies are essential on Chinese made machines or short tube life becomes the norm.

    cheers

    Dave
    You did what !

  4. #4
    I have mine set to 65% in order to stay under 26 mA.
    GWeike LC6090 130 Watt 24" x 36" Table
    Shopbot 96 x 48
    Kossel Clear 3D Printer

  5. #5
    What was your original pump current at 100% Larry?

    cheers

    Dave
    You did what !

  6. #6
    I just adjusted the power supply. I could not access the screw without unscrewing the entire power supply. I had to turn it left by 2.5-3 turns.

    Before:

    10% 4.5 mA
    20% 7.2 mA
    30% 6.5 mA
    40% 10 mA
    50% 15 mA
    60% 20 mA
    70% 23.5 mA
    80% 26.5 mA
    90% 29 mA
    98% ~32 mA

    After:

    10% 4.0 mA
    20% 5.5 mA
    30% 7.0 mA
    40% 7.0 mA
    50% 8.5 mA
    60% 13.2 mA
    70% 18.0 mA
    80% 21.5 mA
    90% 24.0 mA
    98% 26.0 mA

    Notice that before 20 and 30% were about the same (actually 30 was less than 20). After 30 and 40% were about the same. So this is not very linear.

    In any case, I am happy that I can now run at an indicated 100%, even if it is the same as what used to be 80%.
    Triumph TR-9060 with 5200 Chiller and 100 watt Z4 RECI.

  7. #7
    Triumph TR-9060 with 5200 Chiller and 100 watt Z4 RECI.

  8. #8
    I thought about it more and changed it to be 28 mA at 98% (the highest my laser firmware will allow). RECI just says to never run it over 28, so that makes sense to make 28 the max. Now if I set the software to 90%, it will be comfortably at or below 26 mA.

    These things should have hour meters in firmware.
    Triumph TR-9060 with 5200 Chiller and 100 watt Z4 RECI.

  9. #9
    Join Date
    Oct 2004
    Location
    Cape Town, South Africa
    Posts
    3,922
    The shenuis do have hour meters built into their software and board firmware , total time on , total X travel , Y travel and so on.
    There is a kinda dichotomy between "hours" and tube longevity
    8000 hours at an average of 5 hours a day is 1600 days , around 250 working days a year ... around 6 years...
    Will the gas in the tube not have leaked in this time? IE if you had a Reci on the shelf for 7 years - could you expect it to work?
    I'm not too sure of that....
    Rodney Gold, Toker Bros trophies, Cape Town , South Africa :
    Roland 2300 rotary . 3 x ISEL's ..1m x 500mm CnC .
    Tekcel 1200x2400 router , 900 x 600 60w Shenui laser , 1200 x 800 80w Reci tube Shenhui Laser
    6 x longtai lasers 400x600 60w , 1 x longtai 20w fiber
    2x Gravo manual engravers , Roland 540 large format printer/cutter. CLTT setup
    1600mm hot and cold laminator , 3x Dopag resin dispensers , sandblasting setup, acid etcher

  10. #10
    Dave, Not sure what you mean by "pump current"?
    GWeike LC6090 130 Watt 24" x 36" Table
    Shopbot 96 x 48
    Kossel Clear 3D Printer

  11. #11
    Quote Originally Posted by Larry Robinson View Post
    Dave, Not sure what you mean by "pump current"?
    my fault Larry, Lasers are generated when a "Gain" medium is *pumped* either by a DC voltage, an RF signal,a Diode or a Chemical Reaction (leaving out nuclear lasers and other more complex types).

    In ours the gas molecules(CO2 molecules) get "excited" by this pumping and barf out a photon before returning to the unpumped state waiting to be hit again so it can pop out another photon.

    These Photons are in the form of light hence a laser beam.

    In this case it was the pumping current (mA) I was interested in. My fault, sometimes I assume when I shouldn't

    cheers

    Dave
    You did what !

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •